Calculating turbulent fluxes level 0 (uncorrected fluxes)
Level 0 fluxes are uncorrected. They are computed according to the following equations:
Ambient sensible heat flux, uncorrected (H, W m-2)
6‑51
CO2 flux, uncorrected (F0,co2, μmol m-2 s-1)
If CO2 is measured as molar density with an open path analyzer:
6‑52
If CO2 is measured as molar density with a closed path analyzer:
6‑53
If CO2 is measured as mixing ratio:
6‑54
If CO2 is measured as mole fraction:
6‑55
H2O flux, uncorrected (F0,co2, mmol m-2 s-1)
If H2O is measured as molar density with an open path analyzer:
6‑56
If H2O is measured as molar density with a closed path analyzer:
6‑57
If H2O is measured as mixing ratio:
6‑58
If H2O is measured as mole fraction:
6‑59
CH4 flux, uncorrected: (F0,ch4, μmol m-2 s-1)
If CH4 is measured as molar density with an open path analyzer:
6‑60
If CH4 is measured as molar density with a closed path analyzer:
6‑61
If CH4 is measured as mixing ratio:
6‑62
If CH4 is measured as mole fraction:
6‑63
Latent heat flux, uncorrected (LE0, W m-2)
6‑64
Evapotranspiration flux, uncorrected: (E0, kg m-2 s-1)
6‑65
Momentum flux, uncorrected: (T0, kg m-1 s-2)
Calculated after van Dijk et al. (2004), eq. 2.85
6‑66
Friction velocity (u*, ms-1)
Calculated according to its definition as:
6‑67
Potential temperature (Tp, K)
Calculated according to:
6‑68
where P0,a=105 Pa, and is the reference pressure.
Monin-Obukhov length (L, m)
Calculated according to its definition as:
6‑69
where κ = ~0.41, and is the von Kármán constant; g = ~9.81 m s-1, and is the gravity.
Monin-Obukhov stability parameter (ζ, non-dimensional)
ζ is calculated as:
6‑70
where hm is the measurement height above the ground, as measured in the center of the anemometer measurement volume, and d is the displacement height.
Dynamic temperature (T*, K)
Calculated according to its definition (e.g., Foken and Wichura, 1996):
6‑71