Specifications
CO2 measurements
Measurement range: 0 to 20,000 ppm
Accuracy
LI-850: Within 1.5% of reading
LI-830: Within 3% of reading
Calibration drift
Zero drift1: <0.15 ppm / °C
Span drift2: <0.03% / °C
Total drift at 370 ppm3: <0.4 ppm / °C
RMS noise at 370 ppm with 1 sec signal filtering: <1 ppm
Sensitivity to water vapor (LI-850 only): <0.1 ppm CO2 / mmol mol-1 H2O
Lower limit of detection: 1.5 ppm
H2O measurements (LI-850 only)
Measurement range: 0 to 60 mmol mol-1
Accuracy: Better than 1.5% of reading
Calibration drift
Drift at 0 mmol mol-1: <0.005 mmol mol-1 / °C
Span drift at 10 mmol mol-1: <0.006 mmol mol-1 / °C
Total drift at 10 mmol mol-1: <0.016 mmol mol-1 / °C
RMS noise at 10 mmol mol-1 with 1 sec signal filtering: <0.01 mmol mol-1
Sensitivity to CO2: <0.0001 mmol mol-1 H2O / ppm CO2
Lower limit of detection: 0.015 mmol mol-1
General
Output rate: Up to 2 measurements per second
Response time (T90)
CO2: <3.5 seconds from 0 to 375 ppm
H2O (LI-850 only): <3.5 seconds from 0 to 21 mmol mol-1
Measurement principle: Non-Dispersive Infrared
Traceability
CO2: Traceable gases to WMO standards from 0 to 3,000 ppm; traceable gases to EPA protocol gases from 3,000 to 20,000 ppm
H2O (LI-850 only): NIST traceable LI-610 Portable Dew Point Generator
Pressure compensation range: 50 to 110 kPa
Maximum gas flow rate: 1 liter min-1
Output signals: Two analog voltage (0 to 2.5 V or 0 to 5 V) and two current (4 to 20 mA)
Digital outputs: TTL (0 to 5 V) or Open Collector
DAC resolution: 16 bits across user-specified range
Power requirements
Input voltage: 12-30 VDC
After warmup (without pump): 0.33A @ 12 VDC (4.0 W) average
After warmup (with pump): 0.42A @ 12 VDC (5.0 W) average
During warmup: 1.2 A @ 12 VDC (14 W) maximum
Operating temperature range: -20 to 45 °C
Relative humidity range: 0 to 95% RH, Non-condensing
Dimensions: 22.23 cm W × 15.25 cm D × 7.62 cm H
Weight
No pump, no display: 1.0 kg
No pump, with display: 1.02 kg
With pump, no display: 1.3 kg
With pump, with display: 1.32 kg
Internal optical cell volume: 14.5 mL
Pump specifications (optional)
Operating temperature range: 5 to 45 °C
Storage temperature range: -20 to 60 °C
Operating humidity range: 0 to 80% RH
Nominal flow rate: 0.75 liters minute-1
Power consumption: 1 W (nominally)
Expected life span: 8,000 hrs in standard conditions with a normal load
Display specifications (optional)
Dimensions: 6.7 cm corner-to-corner
Resolution: 400 × 200 px; monochrome
Power consumption: <200 µW
Displayed variables: CO2 reading, H2O reading (LI-850 only), optical bench temperature, and pressure.
Dimensional drawings
Serial cable pin assignments
The image below shows the pin assignments for the 9-pin RS-232 serial connector.
Serial communication parameters
The LI-830 and LI-850 can communicate through a RS-232 serial port on the front of the instrument. You may need to set the communication parameters on your computer or your terminal emulator program. The RS-232 port is configured as Data Terminal Equipment (DTE) with no hardware handshaking. It is bi-directional, meaning information can be transferred both into and out of the instrument. The port is configured as follows:
- Baud Rate: 9600 bps
- Data Bits: 8
- Parity: None
- Stop Bits: 1
- Flow Control: None
Specifications subject to change without notice.