Equation Summary

The LI-840A computes CO2 and H2O concentrations using an equation of the form

B‑1

where c is concentration, f() is the calibration function, α is the absorptance, g (α,P) is the pressure correction, S(α) is the span, and T is the temperature (°C) of the gas in the cell, typically 51.5 °C. Absorptance is computed from

B‑2

where V and Vo are the raw detector sample and reference readings, and Z is the zeroing parameter.

Span is a linear function of absorptance.

B‑3

H2O Equations

Absorptance αw for water vapor is computed from

B‑4

where Vw and Vwo are the sample and reference raw detector readings and Zw is the zero parameter. The pressure correction for water vapor is an empirical function gw() of absorptance and pressure P:

B‑5

The value of Po is 99 kPa. When the pressure correction is not enabled, gw() is simply 1.0. Values for gw() can be viewed on the diagnostics window (see Operation).

Water vapor concentration W (mmol mol-1) is computed from

B‑6

where fw(x) is a third order polynomial whose coefficients are given on the calibration sheet.

B‑7

CO2 Equations

The measurement of CO2 is a bit more complicated than for H2O because of the influence of water vapor. There is a slight direct cross sensitivity in the CO2 signal to H2O. This is measured at the factory and accounted for in the computation of absorptance (equation B‑8). There is also a band broadening effect that is accounted for in the computation of concentration (equation B‑12).

CO2 absorptance αc is computed from

B‑8

where Vc and Vco are the raw detector signals for sample and reference, Zc is the CO2 zero parameter, and Xwc is a cross sensitivity parameter for the effect of water vapor on CO2. Its value is reported on the calibration sheet as “XS=”.

The empirical pressure correction function gc() depends on CO2 absorptance and pressure:

When P = Po, gc() = 1.

When P < Po

B‑9

where a = 1.10158, b = -6.1217 * 10-3, c = -0.266278, d = 3.69895, and z is the asymptotic value of absorptance, obtained from the calibration coefficients (equation B‑13).

B‑10

When P > Po

B‑11

where X, A, and B are computed as in equation B‑9. The variable gc() is viewable on the diagnostics window (see Operation).

CO2 concentration C (µmol mol-1) is computed from

B‑12

where fc(x) is a function whose inverse is a double rectangular hyperbola, and whose coefficients (a1…a4) are given on the calibration sheet.

B‑13

Solving equation B‑13 for C yields the calibration function

B‑14

Where

B‑15

ψ(W) accounts for band broadening by water vapor, and is viewable on the Diagnostic window.

B‑16

The band broadening coefficient h(αc) has been determined to be 1.45 for the LI-840A for CO2 concentrations near ambient. At higher concentrations, the value decreases. We capture this behavior with an empirical relationship (equation B‑17).

B‑17

Where z is from equation B‑10, and bw is the low concentration band broadening coefficient: 1.45. This is the value shown on the calibration sheet as BB=1.45. The present value of h(αc) is viewable on the Diagnostics screen (Section 3). The typical relationship between h(αc) and CO2 concentration is illustrated below. (‘Typical’ because the exact relationship depends on the relationship between absorptance and CO2, which is the calibration curve.)

Implementation Note: We formulated B‑17 with 0.64bw – 0.64 instead of the simple equivalent (0.29) because this allows band broadening corrections to be turned off by setting bw to 1. When bw =1, h(αc) = 1 everywhere. Also, to avoid computational problems (underflows, overflows, and division by zero) we constrain the argument αc when computing h(αc) to be

B‑18

is typically equivalent to about 600 ppm.

Calibration Equations

Zeroing H2O

When the command for zeroing water is received, the LI-840A computes the water zero from equation B‑19, where and are averaged for 5 seconds.

B‑19

Zeroing CO2

When the command for zeroing CO2 is received, the LI-840A computes the CO2 zero term from equation B‑20, where , , , and are averaged for 5 seconds.

B‑20

Spanning H2O

When the command for setting the span for H2O is received, along with the target concentration WT, the LI-840A computes Sw0 from equation B‑21 , where is averaged over five seconds.

B‑21

where

B‑22

The instrument retains the following values, which are used for subsequent secondary spans:

B‑23

B‑24

Secondary Span H2O

When the secondary span command for H2O is received, the instrument computes new values for both Sw0 and Sw1. First, it measures a new and computes a new from equation B‑22. Then, it uses these plus the retained values (and from the previous normal span) to compute

B‑25

Given the new span slope Sw1, update the span offset Sw0 by equation B‑21.

Spanning CO2

When the command for setting the span for CO2 is received, along with the target concentration CT, the LI-840A computes Sc0 from equation B‑26, where and are averaged for 5 seconds.

B‑26

where

B‑27

Note that

B‑28

The instrument retains the following values which are used for subsequent secondary spans, if necessary:

B‑29

B‑30

Secondary Span CO2

When the secondary span command for CO2 is received, the instrument computes new values for both Sc0 and Sc1. First, it measures a new and computes a new βc from equation B‑27. Then it uses these, plus the retained values (αc1 and βc1 from the previous normal span) to compute

B‑31

Given the new span slope Sc1, update the span offset Sc0 by equation B‑26.

Symbol Summary

Many of the quantities described in the above section are available from the LI-840A. These are summarized in Table B‑1.

Table B‑1. Symbol summary
Symbol XML Tag: <LI840>+ Description
αw <data><h2oabs> H2O absorptance
gW(αw,P) <AUXdata><pHA> Pressure corrected absorptance of H2O
W <data><h2o> H2O (mmol mol-1)
P <data><cellpress> Pressure (kPa)
T <data><celltemp> Temp (C)
αc <data><co2abs CO2 absorptance
gc(αc,P) <AUXdata><pcA> Pressure corrected absorptance of CO2
C <data><co2> CO2 (µmol mol-1)
Vw <data><raw><h2o>  Raw detector readings
Vwo <data><raw><h2oref>
Vc <data><raw><co2>
Vco <data><raw><co2ref>
aw1...aw3 <poly><h2o> Calibration coefficients
ac1...ac4 <poly><co2> Calibration coefficients
Zw <cal><h2okzero>  
Sw0 <cal><h2okspan>  
Sw1 <cal><h2okspan2>  
Zc <cal><co2kzero> Zero and Span values
Sc0 <cal><co2kspan>  
Sc1 <cal><co2kspan2>  
xwc <poly><xs> Cross sensitivity coefficient
bw <poly><bb> Band broadening coefficient for low CO2
h(αc) <AUXdata><bb_EFF> Computed band broadening coeff for the current absorptance
ѱ(W) <AUXdata><psi> Band broadening correction factor for water vapor