Data dictionary - measured variables
The data dictionary describes variable names from the files and how those variables are displayed in the user interface and available in data files. Each component and sensor can contribute data to the dataset, as described below.
LI-8250 Multiplexer measured variables
The LI-8250 provides data about itself, including status and performance information. Measured variables indicate the performance characteristics and status of the device.
Interface variable | Units | Data file variable | Description |
---|---|---|---|
DOY | DOY | Decimal day of year at start of measurement | |
Date Time | YYYY-MM-DD HH:MM:SS | DATE_TIME | Combined date and time timestamps |
Date | YYYY-MM-DD | DATE | Timestamp of year-month-day |
Time | HH:MM:SS | TIME | Timestamp of hours-minutes-seconds |
Temperature | C | T_CASE | Internal temperature of the multiplexer |
Pressure | kPa | PA | Atmospheric pressure |
Pressure Temperature | C | T_PA | Atmospheric pressure sensor temperature |
Pump Positive Pressure | kPa | PUMP_POSITIVE_PRESSURE | Pressure sensor on the positive pressure side of the sampling pump |
Pump Negative Pressure | kPa | PUMP_NEGATIVE_PRESSURE | Pressure sensor on the negative pressure side of the sampling pump |
Flow | L M-1 | FLOW | Flow rate through the multiplexer to the active port. |
Pump Time | H | PUMP_TIME | Hours of sampling pump operation |
Pump Setpoint | PUMP_SETPOINT | Drive setting for the sampling pump | |
Pump Current | A | PUMP_CURRENT | Current draw by the sampling pump |
Purge Pump Time | H | PURGE_PUMP_TIME | Hours of purge pump operation (purge pump accessory required) |
Purge Pump Setpoint | PURGE_PUMP_SETPOINT | Drive setting for the purge pump (purge pump accessory required) | |
Purge Pump Current | A | PURGE_PUMP_CURRENT | Current draw by the purge pump (purge pump accessory required) |
Purge Pump Pressure | kPa | PURGE_PUMP_PRESSURE | Pressure sensor on the positive pressure side of the purge pump (purge pump accessory required) |
Subsample Pressure | kPa | SUB_PRESSURE | Ambient to subsample loop pressure differential |
Subsample Flow | L M-1 | SUB_FLOW | Flow through the subsample loop |
VSO Flow Setpoint1 | VSO_FLOW_SETPOINT | Setting for the Voltage Sensitive Orifice on the inlet side of the system valve manifold | |
VSO Pressure Setpoint | VSO_PRESSURE_SETPOINT | Setting for Voltage Sensitive Orifice on the outlet side of the system valve manifold | |
Port n Current | A | PORTn_CURRENT | Current draw through the port number n |
Solenoid Current | A | SOL_CURRENT | Current supplied to valves and VSOs |
12 VDC Current | A | 12VDC_CURRENT | Current draw on the 12 VDC power output supply |
Voltage | V | VIN | Input voltage to multiplexer |
Voltage Out | V | VOUT | Output voltage on the nominal 24 VDC supply |
VSO Flow Control Voltage | V | Voltage to the inlet valve on the system valve manifold | |
VSO Pressure Control Voltage | V | Voltage to the outlet side of the system valve manifold | |
5 VDC Voltage | V | 5VDC_VOLTAGE | 5 V supply voltage |
12 VDC Voltage | V | 12VDC_VOLTAGE | 12 V supply voltage |
Flow Pressure | kPa | FLOW_PRESSURE | Pressure measurement used to determine flow rate |
Subsample Flow Pressure | kPa | SUB_FLOW_PRESSURE | Pressure measurement used to determine subsample flow rate |
Latitude | ° | LATITUDE | Latitude reported by GPS |
Longitude | ° | LONGITUDE | Longitude reported by GPS |
Satellites | SATELLITES | Number of satellites | |
Altitude | m | ALTITUDE | Altitude reported by GPS |
8250-01 Extension Manifold measured variables
Measured variables indicate the performance characteristics and status of the device.
Interface variable | Units | Data file variable | Description |
---|---|---|---|
Temperature | C | T_CASE | Internal temperature of the extension manifold |
Pump Time | H | PUMP_TIME | Hours of sampling pump operation |
Pump Setpoint | PUMP_SETPOINT | Drive setting for the sampling pump | |
Pump Current | A | PUMP_CURRENT | Current draw by the sampling pump |
Purge Pump Time | H | PURGE_PUMP_TIME | Hours of purge pump operation (purge pump accessory required) |
Purge Pump Setpoint | PURGE_PUMP_SETPOINT | Drive setting for the purge pump (purge pump accessory required) | |
Purge Pump Current | A | PURGE_PUMP_CURRENT | Current draw by the purge pump (purge pump accessory required) |
Purge Pump Pressure | kPa | PURGE_PUMP_PRESSURE | Pressure sensor on the positive pressure side of the purge pump (purge pump accessory required) |
Port n Current | A | PORTn_CURRENT | Current draw through the port number listed |
Solenoid Current | A | SOL_CURRENT | Current supplied to valves and VSOs |
Voltage | V | VIN | Input voltage to multiplexer |
12 VDC Voltage | V | 12VDC_VOLTAGE | 12 V supply voltage |
5 VDC Voltage | V | 5VDC_VOLTAGE | 5 V supply voltage |
3.3 VDC Voltage | V | 3.3VDC_VOLTAGE | 3.3 V supply voltages |
Diag Code | DIAG | Diagnostic flag bits |
8250-02 Calibration Manifold measured variables
The Calibration Manifold provides additional parameters related to the automated calibration and adjustments.
Interface variable | Units | Data file variable | Description |
---|---|---|---|
Flow | L M-1 | flow | Flow of calibration gas |
VSO | V | vso | Control voltage to Voltage Sensitive Orifice (an internal flow controlling valve) |
VIN | V | vin | Input voltage |
Current | A | isys | Current drawn (amps) |
Long-term Chamber measured variables
The long-term soil chambers (8200-104 and 8200-104C) provide data about themselves. Measured variables indicate the performance characteristics and status of the chamber.
Interface variable | Units | Data file variable | Description |
---|---|---|---|
Temperature | C | TA | Chamber air temperature |
Board Temperature | C | TB | Chamber control board temperature |
Voltage | V | VIN | Chamber input voltage |
Motor Current | A | MOTOR_CURRENT | Current draw by chamber drive motor |
State | STATE | Chamber states are: 1 = closing; 2 = opening; 3 = parking; 4 = manual move; 5 = closed; 6 = open; 7 = parked; 8 = unknown |
|
Diagnostic | DIAG | Encoded diagnostic information |
|
Soil Moisture | m3 m-3 | SWC_1 | Optional soil water content, if you a sensor contributing this data, it well be available in SoilFluxPro. |
Soil Temperature | ° | TS_1 | Optional soil temperature, if you have a sensor contributing this data, it will be available in SoilFluxPro. |
Elapsed Time | s | ETIME | Elapsed time of the measurement. |
LI-870 CO2/H2O Analyzer measured variables
The LI-870 provides gas measurements and data about itself. These parameters indicate the performance characteristics and status of the gas analyzer.
Interface variable | Units | Data file variable | Description |
---|---|---|---|
CO2 | μmol mol-1 | CO2 | Carbon dioxide mole fraction |
CO2 Dry | μmol mol-1 | CO2_DRY | Carbon dioxide dry mixing ratio |
H2O | mmol mol-1 | H2O | Water vapor mole fraction |
Cell Temperature | C | T_CELL | Optical bench temperature |
Cell Pressure | kPa | PA_CELL | Pressure in the optical bench |
CO2 Absorption | CO2_ABS | Absorption by carbon dioxide | |
H2O Absorption | H2O_ABS | Absorption by water vapor | |
Dewpoint | C | DEWPOINT | Dewpoint temperature |
Voltage | V | VIN | Input voltage to the LI-870 |
Flow | L M-1 | FLOW | Flow rate through the LI-870 |
LI-78xx Trace Gas Analyzers measured variables
LI-78xx Trace Gas Analyzers report measurement data and performance information. Measured variables indicate the performance characteristics and status of the Trace Gas Analyzer. Items included depend on the gas analyzer connected.
Interface variable | Units | Data file variable | Description |
---|---|---|---|
Diagnostic | DIAGNOSTIC | Diagnostic value indicating the quality of the data and state of the instrument | |
CO2 Dry | μmol mol-1 | CO2_DRY | Carbon dioxide dry mixing ratio (LI-7810, LI-7815, or LI-7825) |
H2O | mmol mol-1 | H2O | H2O vapor mole fraction |
CH4 Dry | nmol mol-1 | CH4_DRY | CH4 dry mixing ratio (LI-7810 only) |
N2O Dry | nmol mol-1 | N2O_DRY | N2O dry mixing ratio (LI-7820 only) |
16O12C16O | μmol mol-1 | 16O12C16O_DRY | C12 CO2 mixing ratio (LI-7825 only) |
16O13C16O | μmol mol-1 | 16O13C16O_DRY | C13 CO2 mixing ratio (LI-7825 only) |
16O12C18O | μmol mol-1 | 16O12C18O_DRY | O18 CO2 mixing ratio (LI-7825 only) |
16O12C17O | μmol mol-1 | 16O12C17O_DRY | O17 CO2 mixing ratio (LI-7825 only) |
δ13CCO2 | ‰ | CO2C13 | C12:13 isotopic ratio of CO2 (LI-7825 only) |
δ17OCO2 | ‰ | CO2O17 | O16:17 isotopic ratio of CO2 (LI-7825 only) |
δ18OCO2 | ‰ | CO2O18 | O16:18 isotopic ratio of CO2 (LI-7825 only) |
NH3 Dry | nmol mol-1 | NH3_DRY | NH3 dry mixing ratio (LI-7825 only) |
Cavity Temperature | C | T_CAVITY | Optical cavity temperature |
Cavity Pressure | kPa | PA_CAVITY | Optical cavity pressure |
Laser Pressure | kPa | PA_LASER_PHASE | Pressure used to control the laser phase |
Laser Temperature | C | T_LASER | Temperature of the laser |
Residual | RESIDUAL | Residual of absorption spectrum fit | |
Ring Down | μs | RING_DOWN | Ring down time |
Enclosure Temperature | C | T_THERMAL_ENCLOSURE | Temperature of the optical cavity enclosure |
Phase Error | PHASE_ERROR | Error between setpoint and actual laser phase in counts | |
Shift Temperature | C | T_LASER_SHIFT | Difference between factory set laser temperature and operating laser temperature |
Voltage | V | VIN | Input voltage to the LI-78xx |