Specifications
CO2 Measurements
Calibration Range: 0 to 3000 µmol mol-1
Accuracy: Within 1% of reading
Zero Drift (per °C):
±0.1 ppm typical; ±0.3 ppm maximum
RMS Noise (typical @ 370 ppm CO2):
@5 Hz: 0.08 ppm
@10 Hz: 0.11 ppm
@20 Hz: 0.16 ppm
Gain Drift (% of reading per °C @ 370 ppm):
±0.02% typical; ±0.1% maximum
Direct Sensitivity to H2O (mol CO2 mol-1 H2O):
±2.00E-05 typical; ±4.00E-05 maximum
H2O Measurements
Calibration Range: 0 to 60 mmol mol-1
Accuracy: Within 2% of reading
Zero Drift (per °C):
±0.03 mmol mol-1 typical; ±0.05 mmol mol-1 maximum
RMS Noise (typical @ 10 mmol mol-1 H2O):
@5 Hz: 0.0034 mmol mol-1
@10 Hz: 0.0047 mmol mol-1
@20 Hz: 0.0067 mmol mol-1
Gain Drift (% of reading per °C @ 20 mmol mol-1):
±0.15% typical; ±0.30% maximum
Direct Sensitivity to CO2 (mol H2O mol-1 CO2):
±0.02 typical; ±0.05 maximum
General
Fundamental Gas Sampling Rate: 150 Hz
Bandwidth: 5, 10, or 20 Hz; software selectable
Type: Absolute, non-dispersive infrared gas analyzer
Detector: Thermoelectrically cooled lead selenide
Path Length: 12.5 cm (4.72")
Optical Cell Volume: 16 cm3
Optical Cell Thermocouples (Input and Output):
Max. Error: ±0.6 °C over -40 to +50 °C, with maximum departure from reference (block) temperature of ±20 °C
Frequency Response: >10 Hz half power response @ 15 lpm flow rate
Inputs: Ethernet, 4 analog input channels (Bipolar; ±5 V; 16 bit; 300 Hz)
Outputs: RS-232 (20 Hz Maximum); SDM (User selectable to 50 Hz); Ethernet; 6 user-scalable 16 bit DACs (0-5 V range, single-ended)
Differential Pressure Measurements:
Range: ±10 kPa.
Accuracy: 4% of reading
Resolution: 0.3 Pa
Absolute Pressure Sensor:
Range: 65 to 115 kPa absolute
Accuracy: ± 0.1% of full scale
Resolution: 0.002 kPa
Signal Noise (peak-to-peak): 0.002 kPa typical
Operating Temperature Range: -25 to 50 °C (-40 °C verification on request)
Relative Humidity Range: 0-95% (non-condensing)
Data Storage: Removable industrial grade USB flash storage
Data Communication: Ethernet, Synchronous Device for Measurement (SDM; User-selectable to 50 Hz), RS-232 (115200 baud); 20 Samples/Sec. max., 6 DACs (single ended; 0-5 V)
Weatherproof Rating: Tested to IEC IP65 standard
User Interface: Windows® PC software
Power Requirements: 10.5 to 30 VDC
Power Consumption: Typically 30 W during warm-up (maximum 37 W at -40 °C), 8 to 18 W steady state (depending on ambient temperature)
Weight:
Head: 1.8 kg (3.95 lb.)
LI-7550 and Cables: 4.4 kg (10 lb.)
Dimensions:
Head: Diameter 7.5 cm, Length 31 cm
LI-7550 Analyzer Interface Unit: 35 cm × 30 cm × 15 cm
Head Cable: 5 meters. Can be extended to 10 meters
Other Cables: 5 meters
Gas analyzer head dimensions
LI-7550 and Flow Module dimensions (early versions)
The LI-7550 and 7200-101 flow module have the same dimensions. The flow module requires additional clearance for the tube fittings at the top and bottom.
LI-7550 and Flow Module dimensions (new versions)

Flow Module (7200-101 and 7200-102)
Flow Rate Settings: 10-18 SLPM (standard liters per minute) setpoint mode or constant blower drive mode.
Blower Pressure: 3.5 kPa minimum at 100% blower drive at 101.3 kPa and 27 °C ambient.
Accessory Link: UART on RS-485 cable.
Power Requirements:
10.5 - 30 VDC; 50 W Maximum
Typical <15 W @ 25 °C, 15 liters/min. at 0.75 kPa vacuum
Operating Conditions: -20 °C to +50 °C; IEC IP55 specified; Designed for outdoor installation
Size: 35 cm × 30 cm × 15 cm
Weight: 13.7 lbs. (6.2 kg).
Insulated Intake Tube (9972-053)
Outside Diameter: 0.25" (6.35 mm)
Inside Diameter: 0.21" (5.334 mm)
Wall Thickness: 0.020" (0.51 mm)
Length: 40" (101.6 cm)
Intake tube path length when assembled: 102 cm
Intake tube path length with Swagelok® filter: 106.5 cm
Material: 304 Stainless Steel
Heated Intake Tube
Input Voltage: 10.5 - 30 V
Total Output Wattage: 0.1 W to 6 W (Heat density ratio: 2:1 short tube to long tube)
Operating Temperature Range: -40 °C to 50 °C
Data Communication Protocol: RS-485
Intake Tube Inside Diameter: 5.33 mm (0.21”)
Intake Tube Outside Diameter: 6.35 mm (0.25”)
Intake Tube Length: 71.1 cm (28”)
Cable Length: 5 m (16.40’)
Intake Tube Weight: 0.54 kg (1.2 lbs)

SmartFlux 2 System
Data Inputs: Three RS-232/422/485 serial ports
Network Communication: Ethernet (TCP/IP)
Data Storage: USB port for 16 GB industrial grade flash drive
Operating Temperature Range: -25 to 50 °C (-40 to 50 °C verification on request)
GPS Antenna Operating Temperature Range: -30 °C to 80 °C
Relative Humidity Range: 0 to 95% (non-condensing)
Input Voltage: 10 to 30 VDC
Power Consumption: 1.5 watts typical (3.5 watts maximum; excluding pass-through power)
Max Current Passed Through to the Anemometer: 2 A
Dimensions with DIN Clip: 149 x 100 x 37 mm (L×W×H)
Weight with GPS Antenna: 1.32 lb. (0.59 kg)
GPS Antenna Diameter: 61 mm (2.4 in.)
GPS Cable Length: 5 m
GPS Accuracy:
GPS Standard Positioning Service: <15 m; 95% typical
WAAS: <3 m; 95% typical