Variables, data and, diagnostics
This section describes measurements reported by the LI-710 when using the SDI-12 protocol. Fluxes - ET, LE, and H (evapotranspiration, latent energy, and heat) - represent a quantity of something exchanged over a time period, over an area. Evapotranspiration is reported as mm. It is the total (millimeters) water vapor exchanged over the measurement period, over the fetch footprint around the sensor. Other computed parameters - VPD, PA, TA, RH, SVP, and TD (vapor pressure deficit, atmospheric pressure, air temperature, relative humidity, saturation vapor pressure, and dewpoint temperature) are the average for the measurement period. The LI-710 sends up to four packets of results, each including up to nine variables.
Group 0: Results
Group 0 is the main set of measurements, including actual evapotranspiration, a set of measured and computed parameters, the sequence number, and diagnostic information for the measurement period.
| Description | Label | Units | Precision |
|---|---|---|---|
| Actual Evapotranspiration | ET | mm | 3 |
| Latent Energy Flux | LE | W m-2 | 1 |
| Heat Flux | H | W m-2 | 1 |
| Vapor Pressure Deficit | VPD | kPa | 2 |
| Atmospheric Pressure | PA | kPa | 2 |
| Air Temperature (not sonic) | TA | °C | 2 |
| Relative Humidity Ambient | RH | percent | 2 |
| Sequence Number | SEQ | # | 0 |
| Diagnostic Value | DIAG | # (0 - 65535) | 0 |
Group 1: Results and sample count
Group 1 represents results and the total number of samples used in the measurement.
| Description | Label | Units | Precision |
|---|---|---|---|
| Actual Evapotranspiration | ET | mm | 3 |
| Latent Energy Flux | LE | W m-2 | 1 |
| Heat Flux | H | W m-2 | 1 |
| Atmospheric Pressure | PA | kPa | 2 |
| Air Temperature (not sonic) | TA | °C | 2 |
| Relative Humidity Ambient | RH | percent | 2 |
| Sequence Number | SEQ | # | 0 |
| Raw Sample Count | SAMP_CNT | # (up to 36000) | 0 |
| Diagnostic Value | DIAG | # (0 - 65535) | 0 |
Group 2: Air, humidity, and instrument information
Group 2 includes measurements of environmental characteristics without any flux results.
| Description | Label | Units | Precision |
|---|---|---|---|
| Absolute Humidity Ambient | AH | g m-3 | 2 |
| Relative Humidity Ambient | RH | percent | 2 |
| Saturated Vapor Pressure Ambient | SVP | kPa | 2 |
| Vapor Pressure Deficit | VPD | kPa | 2 |
| Atmospheric Pressure | PA | kPa | 2 |
| Air Temperature (not sonic) | TA | °C | 2 |
| Dewpoint Temperature | TD | °C | 2 |
| Tilt | TILT | degrees (0 - 180) | 0 |
Group 3: Performance information and diagnostics
Group 3 includes diagnostic and performance information for the LI-710. The Data QC parameter may be of interest. It indicates how many points were discarded in the processing of the raw data. A high percentage of excluded data indicates that the measurement has poor quality.
| Description | Label | Units | Precision |
|---|---|---|---|
| Pump Voltage | PUMP_V | V | 2 |
| Cell Pressure | PA_CELL | kPa | 2 |
| Cell RH | RH_CELL | % | 2 |
| Cell Temperature | TA_CELL | C | 2 |
| Enclosure RH | RH_ENCL | percent | 2 |
| Flow | FLOW | cm3 min-1 | 0 |
| Input Voltage | INPUT_V | V | 2 |
| Data QC | DATA_QC | percent | 0 |
Group 4: Enhanced analysis variables
Group 4 includes three variable that are used for advanced analysis. These variables are used by LI-COR Cloud when displaying the flux footprint.
| Description | Label | Units | Precision |
|---|---|---|---|
| Standard Deviation of Vertical Wind Component (w) | W_SIGMA | m s-1 | 3 |
| Standard Deviation of Temperature | T_SIGMA | °C | 3 |
| Standard Deviation of Water Vapor Concentration | H2O_SIGMA | mmol mol-1 | 3 |