The list_utility module
The following is a listing of the list_utility.py
file on the LI-6800.
Listing 12‑5. Listing of /home/licor/resources/lib/list_utility.py.
#!/usr/bin/env python3
# -*- coding: utf-8 -*-
# Thanks to Ari Kornfeld @ akorn@carnegiescience.edu
import numpy as np
from random import sample
def linearList(vv1, vv2, nn, rounded=2):
# v1 - starting value,
# v2 - ending value
# n - number of values
v1 = float(vv1)
v2 = float(vv2)
n = int(nn)
v_range = [f for f in list(np.linspace(0, 1, num=n))]
setpoints = [v1 + f*(v2-v1) for f in v_range]
return list(np.around(setpoints, rounded))
def randomList(v1, v2, n, rounded=2):
return sample(linearList(v1, v2, n, rounded=rounded),n)
def makeOrtho(listOfLists, lock_index=-1, max_cor=0.1, outfile=''):
# listOfLists should contain 2 or more lists of setpoints. If sizes not equal, smallest size is n
# returns listOfLists with setpoint sorted to provide maximum orthogonality (minimum correlation)
# if you don't want one of the lists sorted, specify that index as lock_index. 0 = first list, 1 = 2nd, etc.
listCount = len(listOfLists)
n = np.min([len(x) for x in listOfLists]) # n is smallest list size
print(n)
iter = 0
if max_cor < 0.05:
max_cor = 0.05
orthogonal_enough = False
p = lambda i: sample(listOfLists[i], n) if i != lock_index else listOfLists[i][0:n]
while not orthogonal_enough:
result = np.matrix([p(i) for i in range(listCount)])
#to test bad ortho: data.frame(T=T_range, C=C_range, Q = Q_range)
# now check that the variables aren't too strongly correlated
cor1 = np.corrcoef(result);
np.fill_diagonal(cor1, 0.0) # we don't care about self correlation
# print(cor1)
cc = np.max(np.abs(cor1))
# print("cc=", cc)
orthogonal_enough = (cc < max_cor)
iter += 1
if (iter > 500):
max_cor = 0.2 # safety valve
print(iter, "iterations, cc=", cc)
if outfile != '':
try:
file = open(outfile, 'w')
print('corr_coeff=', cc, file=file)
for i in range(n):
line = ''
for j in range(listCount):
line += str(result[j,:].tolist()[0][i])+' '
print(line, file=file)
except Exception as e:
print('Exception in makeOrtho:', str(e))
return [result[i,:].tolist()[0] for i in range(listCount)] # return a list of lists
if __name__ == '__main__':
n=12
t = linearList(25, 15, 3) + linearList(17, 45, n-3) # make temp efficient to work through - no big jumps, and start at ambient.
c = linearList(50, 1000, n, rounded=0) # test size mismatch
q = linearList(20, 2000, n, rounded=0)
(t,c,q) = makeOrtho((t,c,q), lock_index=0, outfile="/Users/jon/out.txt")
print(t)
print(c)
print(q)