Specifications
General
Measurement time:
Porometer: 5 to 15 seconds typically, depending on species, leaf surface characteristics, and leaf conditions
Fluorometer: 1 second
Operating conditions:
Temperature: 0 to 50 °C
Pressure: 50 to 110 kPa
Humidity: 0 to 85%; non-condensing
Weight: 0.68 kg (porometer only); 0.73 kg with fluorometer
Dimensions: 32.4 cm x 16.9 cm x 6.2 cm (L x W x H)
Display:
Dimensions: 6.8 cm diagonally
Resolution: 400 x 200 pixels; sunlight readable monochrome
Keypad: 5-button membrane pad
Battery:
Built-in Li-ion
Operating hours: 8 hours typically
Capacity: 5200 mAh
Recharging time: 3.5 hours typically; 2 hours with Qualcomm® Quick Charge™ 2.0 or 3.0
Data storage: 128 MB
USB specifications:
Communication/charging interface: Micro-B
Qualcomm® Quick Charge™ 2.0 or 3.0 for rapid charging
Universal charging adapter:
Input: 90 to 264 VAC; 50 to 60 Hz
Output: 5 VDC; 1 Amp
Configuration Software: Windows® and macOS® applications
Data Output: Plain text files (comma or tab delimited), and Excel files with embedded equations.
Barcode scanner: 1-D and 2-D, including Code 39, Code 128, PDF417, 100% UPC, Data Matrix, QR Code
Photosynthetically Actively Radiation (PAR) measurement:
Units: Photosynthetic Photon Flux Density (PPFD); μmol m-2 s-1
Calibration accuracy: ±10% of reading; traceable to NIST
Cosine correction: Cosine corrected up to 60° angle of incidence
Porometer
Leaf sizes:
7.5 mm minimum length and width
2.8 mm maximum thickness
Flow rates:
Low: 75 µmol s-1
Medium: 115 µmol s-1
High: 150 µmol s-1
RH sensor accuracy: ±2% RH
Reference temperature: ±0.2 °C
Leaf temperature sensor accuracy: ±0.5 °C
Inlet flow measurement: ±1% of reading from 75 µmol s-1 to 150 µmol s-1
Exhaust flow measurement: ±5% of full scale up to 150 µmol s-1
Parameters:
gsw mol m-2 s-1; gbw mol m-2 s-1; gtw mol m-2 s-1; E mmol m-2 s-1
VPcham kPa; VPref kPa; VPleaf kPa; VPDleaf kPa
H2Oref mmol mol-1; H2Osamp mmol mol-1; H2Oleaf mmol mol-1
Fluorometer
Flash types: User configurable Rectangular and Multiphase Flash™ (MPF)
Measuring light peak wavelength: 625 nm
Measuring light peak intensity:
0 to 10,000 µmol m-2 s-1
Flash intensity: 0 to 7500 µmol m-2 s-1
LED Risk Group: Exempt group in acc. with IEC 62471:2006. The LED does not pose any photobiological hazard.
Parameters:
Fo; Fm; Fv; Fv/Fm; Fs; Fm’; ΦPSII; ETR
GPS Receiver
GPS: Accuracy 2.5 m CEP (circular error probable)