Specifications
LI-190R Quantum Sensor Specifications
Absolute Calibration: ± 5% traceable to the U.S. National Institute of Standards and Technology (NIST)
Sensitivity: Typically 5 μA to 10 μA per 1000 µmol s-1 m-2
Linearity: Maximum deviation of 1% up to 10,000 µmol s-1 m-2
Stability: < ± 2% change over one year
Response Time: Less than 1 μs (2 m cable terminated into a 604 Ω load)
Temperature Dependence: ± 0.15% per °C maximum
Cosine Correction: Cosine corrected up to 82° angle of incidence
Azimuth: < ± 1% error over 360° at 45° elevation
Tilt: No error induced from orientation.
Operating Temperature Range: -40 °C to 65 °C
Detector: High stability silicon photovoltaic detector (blue enhanced)
Sensor Housing: Weatherproof anodized aluminum case with acrylic diffuser and stainless steel hardware; O-ring seal on the sensor base
Size: 2.36 cm Diameter x 3.63 cm (0.93" x 1.43")
Weight: 24 g head; 60 g base and cable (2 m) with screws
Cable Length: 2 m, 5 m, 15 m, 50 m (6.5', 16.4', 49.2', 164')

LI-191R Line Quantum Sensor Specifications
Absolute Calibration: ± 10% traceable to the U.S. National Institute of Standards and Technology (NIST). The LI-191R is calibrated via transfer calibration using a reference LI-190R Quantum Sensor. Transfer error is ± 5% (included in the ± 10%)
Sensitivity: Typically 7 μA per 1000 µmol s-1 m-2
Linearity: Maximum deviation of 1% up to 10,000 µmol s-1 m-2
Response Time: 10 μs.
Temperature Dependence: ± 0.15% per °C maximum
Cosine Correction: Acrylic diffuser
Azimuth: < ± 2% error over 360° at 45° elevation
Operating Temperature Range: -40 °C to 65 °C
Sensitivity variation over length: ± 7% maximum using a 1" wide beam from an incandescent light source
Sensing Area: 1 meter L x 12.7 mm W (39.4" x 0.50")
Detector: High stability silicon photovoltaic detector (blue enhanced)
Sensor Housing: Weatherproof anodized aluminum case with acrylic diffuser and stainless steel hardware
Size: 121.3 L x 2.54 W x 2.54 cm D (47.7" x 1.0" x 1.0")
Weight: 1.4 kg (3.0 lb.)
Cable Length: 2 m, 5 m (6.5', 16.4')
2420 Light Sensor Amplifier Specifications
Power Requirements: +3.8 to 28 V (1 mA over full range)
Operating Temperature Range: -40 °C to 50 °C
Turn-on Time: 120 ms
Amplifier Output: -2.5 to 5.0 V
Output Offset Voltage: -10 to 10 μV
Amplifier Gain Range: 0.025 to 0.375 V μA-1
Amplifier Gain Accuracy: ± 0.1% typical (± 0.15% max) of gain setting
Amplifier Output Noise: 0.5 μV rms (0.375 V μA-1 Gain, 0.1 to 10 Hz Bandwidth)
Amplifier Output Loading: ≥ 10 kΩ
Response Time: 10 Hz
Cable Length: 0.9 m
2420 BL Dimensions

2420 BNC Dimensions

2003S Mounting and Leveling Fixture Dimensions
