Instrument Information
Note: All Users can view the status of instruments at a station.
Under the Stations view, (), click the Information button (
) for the station. Or, click the station name and then Station Information.
The Instruments Table presents a list of instruments at the site.
Click by an instrument to get diagnostic and status information for that instrument.
LI-7500RS and LI-7550 Status

- Hostname: The name of the instrument as it is identified by FluxSuite. Typically this is the serial number, unless it has been changed by a user.
- Posted at: The date and time of the most recent posting.
- Posting next at: The next scheduled posting from the instrument.
USB Status
- Status: The condition of the USB drive and data logging.
- Free: Amount of free space on the USB storage device.
- Size: The total capacity of the USB drive.
Status
- Sensor Head: Indicates either LI-7500A/RS or LI-7200/RS.
- CO2 Signal Strength: The signal strength received by the instrument (0 to 100%). Low signal strength indicates that the lenses should be cleaned.
- Methane Data Sync: If there is an LI-7700 Open Path CH4 Analyzer at the site, this indicates whether the data are synchronized.
- Detector Temperature Regulated: Indicates whether the optical detector is regulated as specified (yes/no).
- Chopper Temperature Regulated: Indicates whether the optical source and chopper housing temperature are regulated (yes/no).
- PLL: Phase Lock Loop. If not OK the instrument may need to be serviced.
LI-7200/RS and LI-7550 Status

- Hostname: The name of the instrument as it is identified by FluxSuite. Typically this is the serial number, unless it has been changed by a user.
- Posted at: The date and time of the most recent posting.
- Posting next at: The next scheduled posting from the instrument.
USB Status
- Status: The condition of the USB drive and data logging.
- Free: Amount of free space on the USB storage device.
- Size: The total capacity of the USB drive.
Status
- Sensor Head: Indicates either LI-7500A/RS or LI-7200/RS.
- CO2 Signal Strength: The signal strength received by the instrument (0 to 100%). Low signal strength indicates that the lenses should be cleaned.
- Inlet Temperature OK: Indicates the status of the upper thermocouple. If not OK, replace the upper thermocouple.
- Outlet Temperature OK: Indicates the status of the lower thermocouple. If not OK, replace the lower thermocouple.
- Methane Data Sync: If there is an LI-7700 Open Path CH4 Analyzer at the site, this indicates whether the data are synchronized.
- Detector Temperature Regulated: Indicates whether the optical detector is regulated as specified (yes/no).
- Chopper Temperature Regulated: Indicates whether the optical source and chopper housing temperature are regulated (yes/no).
- Auxiliary Inputs OK: Indicates status of input signals.
- PLL: Phase Lock Loop. If not OK the instrument may need to be serviced.
- Head Pressure OK: Indicates the status of pressure measurements. Service required if not OK.
7200-101 Flow Module

Flow Module
- Flow Setting (slmp): Indicates the software setting in Standard Liters Per Minute.
- Measured Flow (lpm): Actual flow in Liters Per Minute.
- Measured Flow (slpm): Actual flow in Standard Liters Per Minute.
- Motor Drive (%): The power consumed by the motor (percent of maximum). A high percent may indicate that the filter or inlet needs to be cleaned.
- Pressure Drop (kPa): The pressure reduction resulting from the flow. A high pressure drop indicates that the filter or inlet needs to be cleaned.
7200-050 or 7200-040 Heated Intake Tube

Heated Intake Tube
- Status: Indicates if the heated intake tube is on or off.
- Power Setting (W): Indicates the setting that is configured through the gas analyzer software.0
- Actual Power Applied (W): Indicates the power delivered to the heated intake tube.
- Voltage In (V): The voltage measured by the heated intake tube circuitry. Typically this is the same as the system voltage.
LI-7700 Status

- Hostname: The name of the instrument as it is identified by FluxSuite. Typically this is the serial number, unless it has been changed by a user.
- Signal Strength (RSSI): The Residual Signal Strength Indicator (%).
- Instrument Ready: Indicates that the instrument is powered on and operating as expected.
- Reference Line Locked: Indicates that the instrument is measure a methane gas absorbtion feature. If not locked, you may need to reboot the analyzer or initiate the line lock algorithm.
- Signal OK: Yes if above the signal strength threshold. No if below the threshold, indicating the mirrors may need to be cleaned.
- Spin RPM Target OK: A diagnostic indicating whether the spin motor reaches the target speed.
- Block Temperature Regulated: Yes nominally. If no, service may be required.
- Laser Temperature Regulated: Yes nominally. If no, service may be required.
- Open Path Thermocouple OK: Yes nominally. If No, the thermocouple may need to be replaced.
- Spin Motor: Indicates whether the spin motor was running when the update was submitted.
- Washer Pump: Indicates whether the washer motor was running when the update was submitted.
- Lower Mirror Heater: Indicates the on/off status of the lower mirror heater.
- Upper Mirror Heater: Indicates the on/off status of the lower mirror heater.
- Auxiliary Thermocouple 3, 2, or 1: Indicates whether the thermocouple is working, if they are installed.

No additional information is available at this time. Alerts are provided through the results computed by the SmartFlux System.

No additional information is available at this time. Alerts are provided through the results computed by the SmartFlux System.

No additional information is available at this time. Alerts are provided through the results computed by the SmartFlux System.