Specifications

All specifications are subject to change without notice.

LI-8250 Multiplexer

General

Dimensions: 38.5 cm L × 52 cm W × 18.5 cm H

Weight: 7.7 kg

Sample Volume: 35 cm3

Weatherproof Rating: Tested to IEC IP55 standard

Operating Range

Temperature: -20 to 45 °C

Humidity: 0 to 95% RH, non-condensing

User Data Storage: 8 GB total non-volatile

GPS: Accuracy 2.5 m CEP

Coverage Area

Maximum radius from LI-8250 to chambers: 15.0 m with one cable assembly or 30.0 m with an extension manifold and two cable assemblies

Maximum diameter of measurement circle: 30.0 m with two cable assemblies or 60.0 m with two extension manifolds and four cable assemblies.

Plumbing

Flow rate to/from chambers: ~2 to 3 lpm

Pump type: Diaphragm (pumps in the analyzers subsample air stream in the LI-8250)

Barometric Pressure Sensor

Measurement Range: 20 to 110 kPa

Sensor Accuracy: ±0.4 kPa from 50 to 110 kPa

Resolution: 0.006 kPa

Communication

Seven LED Indicators: Power, Ready, Ethernet activity (3), USB activity (2)

Connectivity: Three Ethernet Ports, Wi-Fi (not available in some countries)

Wi-Fi Compatibility: 2.4 GHz, 802.11 a/b/g/n/ac

Connectivity Ports:

One sealed and strain-relieved USB-A: For connection to LI-870 CO2/H2O Analyzer

Two standard, internal USB-A: For USB mass storage (file transfer) or a country-specific Wi-Fi adapter.

Three sealed and strain-relieved RJ-45 Ethernet: For connection to LI-COR LI-78xx Trace Gas Analyzers, network, or cellular modem. Connectors also accept standard, non-sealed RJ-45 Ethernet cables for lab use.

Output Port Connector: Full-duplex RS-422 communication, 115,200 baud, plus 24 VDC power. Each output port is current-limited to ~1.8 A. Combined steady state power consumption for all chambers should not exceed 1 A. For details about pin assignments see Pin assignments.

Power

Power Requirements: 10 to 30 VDC up to 80 W (120 VAC and 240 VAC when using a power supply). The LI-8250 powers the 8200-104/C chambers, the 8250-01 Extension Manifold, and the LI-870 CO2/H2O Analyzer. Other analyzers are powered separately. See Table F‑1 for total system power requirements.

Output Voltage: 12 VDC, 2 A

Alternate Power Input: 21.9 to 26.5 VDC 80 W

Table F‑1. Total system power requirements.
Typical Power Consumption (W)
Instrument Idle Sampling/Moving Max/Warm-up
LI-8250 4.8 15.6 18.2
8250-01 (each) 0.8 9.2 11.0
8200-104/C (each) 0.36 4.8 N/A
LI-870 5.0 5.0 14.0

Note: The maximum listed for the LI-8250 is a typical maximum, there is no additional warm-up power. Chamber power reaches 4.8 W only when opening or closing, not the entire time of sampling.

Cable and tube assemblies

LI-870 1.2-meter Cable and Tubing assembly (9982-010)

  • Length: 1.2 meters
  • Total Tube Volume: 18.96 cm3
  • Cable Connectors: Sealed power and USB-A-to-USB-B for the LI-870

LI-78xx 1.2-meter Cable and Tubing Assembly (9982-011)

  • Length: 1.2 meters
  • Total Tube Volume: 18.96 cm3
  • Cable Connectors: Sealed RJ-45 network cable for Trace Gas Analyzers

LI-78xx 2-meter Cable and Tubing Assembly (9982-009)

  • Length: 2.0 meters
  • Total Tube Volume: 31.6 cm3
  • Cable Connectors: Sealed RJ-45 network cable for Trace Gas Analyzers

Tube T-split (9982-073)

  • Length: 60 cm total
  • Total Tube Volume: 4.7 cm3

Chamber cable and tubing assembly (9982-056)

  • Length: 15 meters
  • Total Tube Volume: 237 cm3
  • Cable Connectors: Sealed power and data cables for 8200-104/C Long-Term Chambers.

8250-01 Extension Manifold

General

Dimensions: 38.5 cm L × 52 cm W × 18.5 cm H

Weight: 7.4 kg

Sample Volume: 27.9 cm3

Weatherproof Rating: Tested to IEC IP55 standard

Operating Range

Temperature: -20 to 45 °C

Humidity: 0 to 95% RH, non-condensing

Plumbing

Flow rate to/from chambers: ~2 to 3 lpm

Pump type: Diaphragm

Communication

Two LED Indicators: Power, Ready

Input Port Connector: Full-duplex RS-422 communication, 115,200 baud, plus 24 VDC power.

Output Port Connector: Full-duplex RS-422 communication, 115,200 baud, plus 24 VDC power. Each output is current-limited to ~1.1 A. Combined steady state power consumption for all chambers should not exceed 1 A.

Power

Power Requirements: Input voltage range is supplied by LI-8250 Multiplexer. Typical 24V and 1.8A max. See Table F‑2 for 8250-01 power requirements.

Alternate Power Input: 21.9 to 26.5 VDC 80 W

Table F‑2. 8250-01 Extension Manifold power requirements.
Typical Power Consumption (W)
Instrument Idle Sampling/Moving Max/Warm-up
8250-01 (each) 0.8 9.2 11.0
8200-104/C (each) 0.36 4.8 N/A

Note: The maximum listed for the 8250-01 is a typical maximum, there is no additional warm-up power. Chamber power reaches 4.8W only when opening or closing, not the entire time of sampling.

Indoor AC to DC Power Supply (8250-772)

Dimensions: 14.5 cm L x 6 cm W x 3.2 cm H

Weight: 0.5 kg

Power requirements: 100 to 240 VAC, 1.3 A, 50/60 Hz, 120 VA

Output Voltage: 24 VDC, 3.75 A

Outdoor AC to DC Power Supply (8250-770)

Dimensions: 15 cm L × 12.5 cm W × 8.5 cm H

Weight: 1.3 kg

Weatherproof Rating: Tested to IEC IP55 standard

Environmental Operating Conditions: Tested to IP55 for dust and water ingress

Operating Range: -20 to 50 °C

Power Requirements:

Serial number 82P-0364 and earlier: 100 to 240 VAC, 50/60 Hz, 120 VA

Serial number 82P-0365 and later: 100 to 240 VAC; 50/60 Hz, 140 VA

Output Voltage:

Serial number 82P-0364 and earlier: 12 VDC, 6.67 A

Serial number 82P-0365 and later: 12 VDC, 8.0 A

Opaque Long-Term Chamber (8200-104)

Dimensions: 48.3 cm L × 38.1 cm W × 33.0 cm H

Weight: 7.3 kg

Chamber Volume: 3955 cm3

Soil Area Exposed: 317.8 cm2

Weatherproof Rating: Tested to IEC IP55 standard

Air Temperature Thermistor

Operating Range: -20 to 50 °C

Accuracy: ±0.3 °C from -20 to 50 °C

Light Sensor Current Input

Range: 0 to 90 µA

Resolution: 1.5 nA

Accuracy: ±(0.37 % of Reading + 8 nA) from -20 to 50 °C

SDI-12 Communications Interface

Max Number of Devices: 10

Output Voltage: 12 VDC, 200 mA

Clear Long-Term Chamber (8200-104C)

Dimensions: 48.3 cm L × 38.1 cm W × 33.0 cm H

Weight: 7.3 kg

Chamber Volume: 3955 cm3

Soil Area Exposed: 317.8 cm2

Weatherproof Rating: Tested to IEC IP55 standard

Air Temperature Thermistor

Operating Range: -20 to 50 °C

Accuracy: ±0.3 °C from -20 to 50 °C

Light Sensor Current Input

Range: 0 to 90 µA

Resolution: 1.5 nA

Accuracy: ±(0.37 % of Reading + 8 nA) from -20 to 50 °C

SDI-12 Communications Interface

Max Number of Devices: 10

Output Voltage: 12 VDC, 200 mA

Flask Sampling Kit (8250-660)

Power

Power Requirements: Input voltage is supplied by the LI-8250 Multiplexer or the 8250-01 Extension Manifold. When a Purge Pump is configured to an extension manifold, the extension manifold must be connected to a power supply, such as the Indoor Power Supply (8250-772), via the Alternate Power Input. See Table F‑3 for 8250-660 power requirements.

Table F‑3. 8250-660 Flask Sampling Kit power requirements.
Typical Power Consumption (W)
Instrument Idle Sampling/Moving Max/Warm-up
9982-079 Thermistor Input Module (each) 0.12 0.12 N/A
9982-077 Purge Pump (each) 5.8 5.8 7.0
Flask Sampling Kit Total (each)   5.92 5.92 7.12

Note: The power requirements for the 9982-077 Purge Pump are based on the pump being configured to run in the user interface. If the pump is not configured to run, it will use no power.

Thermistor

Thermistor Input Module (9982-079)

Operating Range: -20 to 70 °C

Optional Thermistor Assembly (9982-081)

Operating Range: -20 to 70 °C

Accuracy: ±0.3 °C from -20 to 70 °C