Initial assembly

The following section covers the basic steps you will follow to assemble the LI-7200RS for the first time. You'll need the analyzer head, LI-7550, flow module, SmartFlux 2 System, and the accessories.

Preparing the gas analyzer

Begin by installing tubing and data cables that connect the head, flow module, and LI-7550. Also install the LI-7550 USB drive.

Installing the flow module tubing

The 5 m (16.4’) length of Tygon tubing is used for air flow between the LI-7200RS head and the flow module. To attach tubing:

  1. Install the hose barb on the head outlet.
  2. On the LI-7200RS head, remove the compression fitting on the AIR OUT port and install the ¼” hose barb (part number 300-11180 from the flow module spares kit). Tighten the hose barb with an open end wrench.
  3. Remove the red plugs from the EXHAUST port on the connector panel and the port on the top of the case of the flow module.
  4. Install the inlet hose barb fitting (part number 300-11179) in the port on the top of the case.
  5. Tighten the fitting securely with an open end wrench.
  6. Install the exhaust fitting (6572-047 for the 7200-101) or cylindrical exhaust screen (301-18828 for the 7200-102) in the exhaust port.
  7. The screen keeps bugs out of the flow module outlet. Tighten the fitting securely on the exhaust outlet. If using a muffler instead, see Adding a muffler to the air outlet.

  1. Install the inlet hosebarb on the flow module.
  2. The hose barb fits into the bottom of the 7200-101 or top of 7200-102.

  1. Install the black Tygon tubing between the LI-7200RS head air outlet and the flow module air inlet.

  1. Connect the tubing at the hose barb on the LI-7200RS sensor head with a hose clamp from the spares kit. Connect the other end to the flow module inlet.
  2. After you install the analyzer and flow module at the site, trim the tube to a suitable length. The appropriate length for the tube will depend on your site configuration - the distance between the analyzer head and the flow module.

Important: Always use hose clamps to secure the Tygon tubing to the hose barbs. The tubing will initially seal without hose clamps, but leaks will develop slowly over time if the tube is not secured with hose clamps.

Attaching the intake tube

The LI-72000RS includes an insulated 1-meter intake tube (part number 9972-053). The insulated intake tube can be shortened, which, can reduce temperature attenuation, but make it difficult to mount the analyzer without distorting the wind flow. A longer intake tube can lead to higher water vapor signal attenuation and require more power from the pump. The ideal length is between 0.5 and 1 meter, but you can make it as short as 15 cm.

Leave the insulation intact to prevent condensation from forming inside the tube. A reflective film can be placed on top of the insulation to minimize solar heating of the intake surface near the sonic anemometer.

An optional heated intake tube (part number 7200-050) is recommended in extremely cold or humid (>50% RH) environments. The heated intake tube cannot be shortened or bent.

Figure 2‑1. Install either the insulated intake tube (left) or heated intake tube (right) on the air inlet using a compression fitting with a ferrule.

The heated intake tube cable connects to the AUXILIARY connector on the LI-7550 using the auxiliary splitter cable (part number 392-17153), as shown in Figure 2‑2. Contact LI-COR if you want to use the heated intake tube on an LI-7550 manufactured before June 6, 2014 (serial numbers AIU-1279 and lower)

Figure 2‑2. The auxiliary splitter cable should be used when there is both a flow module and a heated intake tube. Older versions of the splitter are T-shaped.

Installing the 2-micron particulate filter

An optional 2-micron Swagelok® filter (part number 9972-073) can be placed in line on the inlet tube. It will reduce the amount of dust that enters the gas analyzer optical cell. Under normal conditions, the filter should extend the amount of time that can pass before you need to clean the optical cell. The filter is recommended in environments that have airborne dust and pollen that can contaminate the optical cell.

Important: The filter does not eliminate the need to regularly clean the optics. Regular cleaning is required in order to prevent measurement drift that can occur when certain kinds of contaminants accumulate on the cell windows.

  • The filter will increase the power requirements of your system. As the filter becomes dirty, the flow module will use more power to maintain the same flow rate.
  • The filter may affect H2O frequency response, especially when it is dirty. This can be addressed with additional frequency response corrections during data processing, but it is preferable to operate the system with a clean filter.
  • The effectiveness of the filter depends upon the site conditions, overall level of dust in the air, and the characteristics of the dust (fine vs. coarse dust particles).
Figure 2‑3. On the insulated intake tube, the filter is added to the end of the tube, followed by an additional piece of tubing (2" by ¼").
Figure 2‑4. On the heated intake tube, it is installed in place of the spacer. Loosen both nuts and remove the spacer. Leave both nuts and ferrules in place; tighten them onto the filter.

Installing the air inlet

The air intake kit includes a flexible air inlet, screen, and hose clamp. It also has a spare screen and hose clamp. The intake cap attaches to either the insulated intake tube or the heated intake tube. If using the insulated intake tube with the particulate filter, install the tube extension to the intake side of the filter. To install the intake cap, insert it into the hose clamp, expand the clamp with a pliers, then slip the air inlet over the tube. The tube will insert about 2.5 cm into the inlet.

Note: Do not expand the hose clamp more than necessary. Doing so can permanently deform the clamp.

Connecting the analyzer data cables

The cable bundle has two cables and a tube. The cable with a yellow connector and a 90° bend connects to the head between the air inlet and outlet; the other end connects to the SENSOR connector on the LI-7550. The cable with stainless steel collars connects to the bottom of the head; the other end connects to the IRGA connector on the LI-7550. Align the notches on the cable connectors, then push in and turn the connector clockwise until it is tight. Continue to push in while tightening the connector until the connector is fully seated. The gasket must be compressed to ensure a watertight seal. The tube attaches to the hose barb between the air inlet and outlet on the head. The other end connects to the quick-connect labeled PRESSURE on the LI-7550. See Figure 2‑5 for details.

Figure 2‑5. The analyzer head cable bundle has two cables and a tube. Connect the cables and tube to both the head and LI-7550 as shown. Press the connectors together while tightening until the connector is fully seated.

Connecting the flow module data cable

The 5-meter accessory cable (part number 392-10092) has male and female connectors that attach to the flow module and LI-7550 connector panels at the connectors labeled ACCESSORY.

Figure 2‑6. The flow module data cable connects the connectors labeled ACCESSORY on the flow module and the LI-7550.

Installing the LI-7550 USB drive

Open the cover to the LI-7550 and insert the USB drive from the spares kit into the USB port. Close and latch the cover.

Important: A USB drive must be installed in the LI-7550 to log data. Beginning with SmartFlux firmware v2.2.8, a USB drive must also be installed in the SmartFlux 2 System to ensure successful data processing. A full dataset is stored on both USB drives.

Preparing the enclosure

Start by assembling the enclosure hardware.

Installing the system power supply cable

Connect the main power supply conduit to the small opening in the enclosure. Orient the lock-nut teeth so that they make contact with the enclosure surface. Connect the main power supply wires to terminals 3, 4 and 5.

Connect the main power supply conduit to the enclosure and attach each wire to the proper terminal - red to red, green to green, and black to black..

As a good practice, be sure that your power cable is not connected to a power source until the wiring is complete. Keep the breakers in the OFF position until you are ready to power up the system.

Table 2‑1. Power wire connections for the main power supply to the enclosure.
From Wire Color To
Main power (-) black DIN Terminal 3 (bottom)
Earth ground green DIN Terminal 4 (bottom)
Main power (+) red DIN Terminal 5 (bottom)

In both the eddy covariance system enclosure (7900-050) and the Biomet Data Acquisition System enclosure (7900-126), install the three strain-relief couplings in the three bottom openings.

Attach the locknuts to the fittings

In the eddy covariance system enclosure (7900-050), install the Power Distribution Kit (7900-235) on the raised DIN rail. Connect the red wire lead between the top of the breaker and top right red terminal. Install one shorting block to connect all black terminals and another to connect all red terminals.

Power wires attach to the connections as shown.
Figure 2‑7. The power distribution kit (7900-235) installed in the enclosure (7900-050).

Preparing the SmartFlux System and network switch

Follow the steps below when installing the SmartFlux System in the 7900-050 Eddy Covariance System Enclosure or 7900-126 Biomet Data Acquisition System Enclosure. If you are using a different enclosure, connect the power and data cables similarly to what is described here.

Install the USB drive

Data are recorded to the USB drive. Install the USB drive in the USB port on the SmartFlux System.

Insert a USB storage device into the SmartFlux System

Connect the GPS antenna cable

Route the GPS cable through the left opening at the bottom of the enclosure and plug it into the GPS port on the SmartFlux System.

Connect the GPS antenna cableConnect the GPS antenna cable

Connect power wires and the Ethernet cable to the SmartFlux System

Connect wire leads to the positive and negative terminals on the SmartFlux System. The black lead connects to negative (-), the red lead connects to positive (+). Plug in the 30 cm Ethernet cable that came with the SmartFlux System. There are two connections for positive (+) and two for negative (-). Use only one pair of +/- terminals.

Connect the power wires

Connect power wires to the Brainboxes network switch

Power wires connect to the positive and negative terminals on the switch. The black lead connects to terminal 1 (-), the red lead connects to terminal 2 (+).

Brainboxes power wires

Installing the SmartFlux System and network switch

Install the SmartFlux 2 or 3 System onto left side of the upper-right DIN rail. Insert the spring-loaded clip on the bottom edge of the DIN rail and compress the spring. Rotate it into place. Mount the switch next to it.

Mount the SmartFlux SystemInstall the network switch
Biomet enclosure assembled

Connecting component power wires

The 7900-050 and 7900-126 enclosures present different wiring options. Be sure the power supplied to the enclosure is within the upper voltage limits of attached components, but capable of providing 4 to 5 amps. Keep the breaker OFF while connecting power wires.

7900-050 Eddy Covariance System Enclosure

With the 7900-050 enclosure, keep the breaker OFF. Connect the positive (+) leads to the red terminals and the negative leads (-) to the black terminals for all components of the system. Proceed to Connecting gas analyzer and flow module power wires.

7900-126 Biomet Data Acquisition System Enclosure

With the 7900-126 enclosure, each component is powered individually from the power supply terminals, the DC-DC converter, or the DRM power outputs. 24-volt and 12-volt power supplies will be wired differently.

Caution: Do not short circuit the power supplies. For example, avoid connecting the V4 POWER OUT terminals from a Data Retention Module to the Power In terminals on a device that is powered from another power supply. Doing so will damage components and may present risk of a fire.

Select the power supply option that describes your system:

24 VDC power supply with Data Retention Module

12 VDC power supply with Data Retention Module

12 VDC power supply with no Data Retention Module

24 VDC power supply with the TDK-Lambda DC-DC converter

Connecting gas analyzer and flow module power wires

Power wires for the gas analyzer and flow module connect to terminals in the enclosure. With the eddy covariance system enclosure (7900-050) and Power Distribution Kit (7900-235), connect the red and black power wires leads to corresponding red and black terminals. With the Biomet Data Acquisition System enclosure (7900-126), connect the leads as described in Table 2‑6.

Table 2‑6. Power wire connections for the Biomet Data Acquisition System.
From Wire Color To
Terminal 17 (bottom) black LI-7550 Power (-)
Terminal 19 (bottom) red LI-7550 Power (+)
Terminal 18 (bottom) black Flow Module Power (-)
Terminal 20 (bottom) red Flow Module Power (+)

Connecting Ethernet cables

Ethernet cables connect the SmartFlux 2 System and LI-7550 to the network switch. Vacant ports on the network switch are for other networked components, such as the cellular gateway, PhenoCam, LI-7700, or connecting a PC.

Figure 2‑11. Install Ethernet cables to connect the SmartFlux 2 System and LI-7550 to the Network Switch.

Connecting sonic anemometer data and power cables

The SmartFlux 2 System can record digital data from a number of sonic anemometer models. These are described below. Power for the sonic anemometer is provided through the SmartFlux System.

Gill WindMaster or WindMaster Pro

Gill HS-50 or R3-50

Campbell Scientific CSAT3

Campbell Scientific CSAT3B

See CSAT3B and CSAT3 for details on the USB connection.

Metek Cage or Class-A

RM Young 81000x